Laser absorption spectrometry
Encyclopedia
Laser absorption spectrometry (LAS) refers to techniques that use lasers to assess the concentration or amount of a species in gas phase by absorption spectrometry (AS).
Optical spectroscopic techniques in general, and laser-based techniques in particular, have a great potential for detection and monitoring of constituents in gas phase. They combine a number of important properties, e.g. a high sensitivity and a high selectivity with non-intrusive and remote sensing capabilities. Laser absorption spectrometry has become the foremost used technique for quantitative assessments of atoms and molecules in gas phase. It is also a widely used technique for a variety of other applications, e.g. within the field of optical frequency metrology or in studies of light matter interactions. The most common technique is Tunable Diode Laser Absorption Spectrometry (TDLAS) which has become commercialized and is used for a variety of applications.
There are basically three ways to improve on the situation:
The first can be achieved by the use of a modulation technique; the second can be obtained by using transitions in unconventional wavelength regions, whereas the third by using external cavities.
Due to their good tunability and long lifetime (> 10 000 hours), most practical laser-based AS is today performed by distributed feedback (DFB) diode lasers. This gives rise to systems that can run unattended for thousands of hours, with a minimum of maintenance.
However, since these laser are mostly developed for the telecom industry, they emit in the near infrared (NIR) region, primarily in the 700 nm – 2 μm range. With light in this wavelength region, mostly only weak overtone transitions of molecules can be addressed. This limits the detectability of conventional TDLAS to detection of species down to the mid or high ppm m range (part-per-million concentrations times meter interaction lengths). This is still insufficient for a large range of applications, wherefore other actions have to be taken.
The recent development of quantum cascade laser
s (QC) lasers working in the MIR region has opened up new possibilities for sensitive detection of molecular species on their fundamental vibrational bands. It is more difficult (although not impossible) to generate stable cw light addressing electronic transitions, since these often lie in the UV region. Despite the fact that transitions with larger transition strength can be reached by either MIR or UV emitting lasers, the limits of detection (LODs) of AS techniques using these transitions have not yet been improved as much as has been anticipated. The reason is that these types of laser have a number of unique properties that limit their practical applicability. Their full potential can only be used whence these limitations have been circumvented. This is a rapidly developing but still only partly explored field of science that can help overcoming some of the present limitations of the AS technique.
External cavities can either be of multi-pass type, i.e. Herriott or White cells
, or be of resonant type, most often working as a Fabry–Pérot (FP) etalon. Whereas the multi-pass cells typically can provide an enhanced interaction length of up to ~2 orders of magnitude, the resonant cavities can provide a much larger path length enhancement, in the order of the finesse of the cavity, F, which for a balanced cavity with high reflecting mirrors with reflectivities of ~99.99–99.999% can be ~ 104 to 105. It should be clear that if all this increase in interaction length can be used efficiently, this vouches for a significant increase in detectability!
A problem with resonant cavities is though that a high finesse cavity has very narrow cavity modes, often in the low kHz range (the width of the cavity modes is given by FSR/F, where FSR is the free-spectral range of the cavity, which is given by c/2L, where c is the speed of light and L is the cavity length). Since cw lasers often have free-running linewidths in the MHz range, and pulsed even larger, it is non-trivial to couple laser light effectively into a high finesse cavity. There are though a few ways this can be achieved.
(NICE-OHMS). The first and so far ultimate realization of this technique, performed for frequency standard applications, reached an astonishing LODs of 5•10−13 (1•10−14 cm−1). It is clear that this technique, correctly developed, has a larger potential than any other technique for trace gas analysis!
Optical spectroscopic techniques in general, and laser-based techniques in particular, have a great potential for detection and monitoring of constituents in gas phase. They combine a number of important properties, e.g. a high sensitivity and a high selectivity with non-intrusive and remote sensing capabilities. Laser absorption spectrometry has become the foremost used technique for quantitative assessments of atoms and molecules in gas phase. It is also a widely used technique for a variety of other applications, e.g. within the field of optical frequency metrology or in studies of light matter interactions. The most common technique is Tunable Diode Laser Absorption Spectrometry (TDLAS) which has become commercialized and is used for a variety of applications.
Direct Laser Absorption Spectrometry (DLAS)
The most appealing advantages of LAS is its ability to provide absolute quantitative assessments of species. Its biggest disadvantage is that it relies on a measurement of a small change in power from a high level; any noise introduced by the light source or the transmission through the optical system will deteriorate the detectability of the technique. Direct laser absorption spectrometric (DLAS) techniques are therefore often limited to detection of absorbance ~10−3, which is far away from the theoretical shot noise level, which for a single pass DAS technique is in the 10−7 – 10−8 range. This detectability is insufficient for many types of applications.There are basically three ways to improve on the situation:
- to reduce the noise;
- to address transitions with larger transitions strengths and
- to increase the interaction length.
The first can be achieved by the use of a modulation technique; the second can be obtained by using transitions in unconventional wavelength regions, whereas the third by using external cavities.
Modulated techniques – WMS and FMS – Tunable Diode Laser Absorption Spectrometry (TDLAS)
Modulation techniques make use of the fact that technical noise usually decreases with increasing frequency (often referred to as a 1/f noise) and improves on the signal contrast by encoding and detecting the absorption signal at a high frequency, where the noise level is low. The most common modulation techniques, wavelength modulation spectroscopy (WMS) and frequency modulation spectroscopy (FMS), achieve this by rapidly scanning the frequency of the light across the absorbing transition. Both techniques have the advantage that the demodulated signal is (ideally) zero in the absence of absorbers but they are also limited by residual amplitude modulation (RAM), either from the laser or from multiple reflections in the optical system (etalon effects). The most frequently used laser-based technique for environmental investigations and process control applications is based upon diode lasers and WMS and often referred to as Tunable Diode Laser Absorption Spectroscopy (TDLAS). The typical detectability of WMS and FMS techniques is in the 10−5 range.Due to their good tunability and long lifetime (> 10 000 hours), most practical laser-based AS is today performed by distributed feedback (DFB) diode lasers. This gives rise to systems that can run unattended for thousands of hours, with a minimum of maintenance.
However, since these laser are mostly developed for the telecom industry, they emit in the near infrared (NIR) region, primarily in the 700 nm – 2 μm range. With light in this wavelength region, mostly only weak overtone transitions of molecules can be addressed. This limits the detectability of conventional TDLAS to detection of species down to the mid or high ppm m range (part-per-million concentrations times meter interaction lengths). This is still insufficient for a large range of applications, wherefore other actions have to be taken.
Laser absorption spectrometry using fundamental vibrational or electronic transitions
The second way of improving the detectability of LAS is to employ transitions with larger linestrength, either in the fundamental vibrational band or electronic transitions. The former, which normally reside at ~5 μm, have linestrengths that are ~2–3 orders of magnitude higher than those of typical overtone transition. As an example, the figure (to appear soon) shows the linestrengths for the second and first overtones in NO (at ~1.8 and 2.65 µm) as well as those for the fundamental vibrational transitions in the MIR region (around 5.3 µm). On the other hand, electronic transitions have often yet another 1–2 orders of magnitude larger linestrengths. The transitions strengths for the electronic transitions of NO, which are located in the UV range (at ~227 nm, not included in the figure) are ~2 orders of magnitude larger than those in the MIR region! If transitions with such large linestrengths can be used efficiently, a significant increase in detectability would result.The recent development of quantum cascade laser
Quantum cascade laser
Quantum cascade lasers are semiconductor lasers that emit in the mid- to far-infrared portion of the electromagnetic spectrum and were first demonstrated by Jerome Faist, Federico Capasso, Deborah Sivco, Carlo Sirtori, Albert Hutchinson, and Alfred Cho at Bell Laboratories in 1994.Unlike typical...
s (QC) lasers working in the MIR region has opened up new possibilities for sensitive detection of molecular species on their fundamental vibrational bands. It is more difficult (although not impossible) to generate stable cw light addressing electronic transitions, since these often lie in the UV region. Despite the fact that transitions with larger transition strength can be reached by either MIR or UV emitting lasers, the limits of detection (LODs) of AS techniques using these transitions have not yet been improved as much as has been anticipated. The reason is that these types of laser have a number of unique properties that limit their practical applicability. Their full potential can only be used whence these limitations have been circumvented. This is a rapidly developing but still only partly explored field of science that can help overcoming some of the present limitations of the AS technique.
Cavity Enhanced Absorption Spectrometry (CEAS)
The third way of improving the detectability of LAS is to extend the interaction length. This can be obtained by placing the species inside a cavity in which the light bounces back and forth many times, whereby the interaction length can be increased considerably. This has led to a group of techniques denoted as cavity enhanced AS (CEAS). The cavity can either be placed inside the laser, giving rise to intracavity AS, or outside, when it is referred to as an external cavity. Although the former technique can provide a high sensitivity, its practical applicability is limited because of all the non-linear processes involved.External cavities can either be of multi-pass type, i.e. Herriott or White cells
White cell (spectroscopy)
A White cell is a type of long path gas phase spectroscopy cell that is commonly used to measure low-concentration components in gases or liquids. In 1942 John U...
, or be of resonant type, most often working as a Fabry–Pérot (FP) etalon. Whereas the multi-pass cells typically can provide an enhanced interaction length of up to ~2 orders of magnitude, the resonant cavities can provide a much larger path length enhancement, in the order of the finesse of the cavity, F, which for a balanced cavity with high reflecting mirrors with reflectivities of ~99.99–99.999% can be ~ 104 to 105. It should be clear that if all this increase in interaction length can be used efficiently, this vouches for a significant increase in detectability!
A problem with resonant cavities is though that a high finesse cavity has very narrow cavity modes, often in the low kHz range (the width of the cavity modes is given by FSR/F, where FSR is the free-spectral range of the cavity, which is given by c/2L, where c is the speed of light and L is the cavity length). Since cw lasers often have free-running linewidths in the MHz range, and pulsed even larger, it is non-trivial to couple laser light effectively into a high finesse cavity. There are though a few ways this can be achieved.
Cavity Ring-Down Spectrometry (CRDS)
In cavity ring-down spectrometry (CRDS) the mode-matching condition is circumvented by injecting a short light pulse in the cavity. The absorbance is assessed by comparing the cavity decay times of the pulse as it “leaks out” of the cavity on and off-resonance, respectively. While independent of laser amplitude noise, this technique is often limited by drifts in the system between two consecutive measurements and a low transmission through the cavity. Despite this, sensitivities in the ~10-7 range can routinely be obtained (although the most complex setups can reach below this~10-9). CRDS has therefore started to become a standard technique for sensitive trace gas analysis under a variety of conditions.Also crds is now very effective method for different physical parameters(such as temperature,pressure,strain)sensing.Integrated Cavity Output Spectroscopy (ICOS)
Integrated cavity output spectroscopy (ICOS) sometimes called as cavity-enhanced absorption spectroscopy (CEAS) records the integrated intensity behind one of the cavity mirror, while the laser is repeatedly swept across one or several cavity modes. However, for high finesse cavities the ratio of “on” and “off” a cavity mode is small, given by the inverse of the finesse, whereby the transmission as well as the integrated absorption becomes small. Off-axis ICOS (OA-ICOS) improves on this by coupling the laser light into the cavity from an angle with respect to the main axis so as do not interact with a high density of transverse modes. Although intensity fluctuations are lower than direct on-axis ICOS, the technique is, however, still limited by a low transmission and intensity fluctuations due to partly excitation of high order transverse modes, and can again typically reach sensitivities ~10−7 .Continuous wave Cavity Enhanced Absorption Spectrometry (cw-CEAS)
The group of CEAS techniques that has the largest potential to improve is that based on a continuous coupling of laser light into the cavity. This requires however an active locking the laser to one of the cavity modes. There are two ways in which this can be done, either by optical or electronic feedback,. Optical feedback (OF) locking, originally developed by Romanini et al. for cw-CRDS, uses the optical feedback from the cavity to lock the laser to the cavity while the laser is slowly scanned across the profile (OF-CEAS). In this case, the cavity needs to have a V-shape in order to avoid OF from the incoupling mirror. OF-CEAS is capable of reaching sensitivities ~10−8 range, limited by a fluctuating feedback efficiency. Electronic locking is usually realized with the Pound-Drever-Hall (PDH) technique, and is nowadays a well established technique, although it can be difficult to achieve for some types of lasers. It has been shown by that also electronically locked CEAS can be used for sensitive AS on overtone lines.Noise-Immune Cavity-Enhanced Optical-Heterodyne Molecular Spectroscopy (NICE-OHMS)
However, all attempts to directly combine CEAS with a locking approach (DCEAS) have one thing in common; they do not manage to use the full power of the cavity, i.e. to reach LODs close to the (multi-pass) shot-noise level, which is roughly 2F/π times below that of DAS and can be down to ~10−13. The reason is twofold: (i) any remaining frequency noise of the laser relative to the cavity mode will, due to the narrow cavity mode, be directly converted to amplitude noise in the transmitted light, thereby impairing the detectability; and (ii) none of these techniques makes use of any modulation technique, wherefore they still suffers from the 1/f noise in the system. There is, however, one technique that so far has succeeded in making full use of the cavity by combining locked CEAS with FMS so as to circumvent both of these problems, and that is Noise-Immune Cavity-Enhanced Optical-Heterodyne Molecular SpectroscopyNoise-Immune Cavity-Enhanced Optical-Heterodyne Molecular Spectroscopy
Noise-immune cavity-enhanced optical-heterodyne molecular spectroscopy is an ultra-sensitive laser-based absorption technique that utilizes laser light to assess the concentration or the amount of a species in gas phase by absorption spectrometry .- Principles :The NICE-OHMS technique combines...
(NICE-OHMS). The first and so far ultimate realization of this technique, performed for frequency standard applications, reached an astonishing LODs of 5•10−13 (1•10−14 cm−1). It is clear that this technique, correctly developed, has a larger potential than any other technique for trace gas analysis!
See also
- Absorption spectroscopyAbsorption spectroscopyAbsorption spectroscopy refers to spectroscopic techniques that measure the absorption of radiation, as a function of frequency or wavelength, due to its interaction with a sample. The sample absorbs energy, i.e., photons, from the radiating field. The intensity of the absorption varies as a...
- Cavity Ring Down Spectroscopy (CRDS)Cavity ring down spectroscopyCavity ring-down spectroscopy is a highly sensitive optical spectroscopic technique that enables measurement of absolute optical extinction by samples that scatter and absorb light. It has been widely used to study gaseous samples which absorb light at specific wavelengths, and in turn to...
- Diode LasersLaser diodeThe laser diode is a laser where the active medium is a semiconductor similar to that found in a light-emitting diode. The most common type of laser diode is formed from a p-n junction and powered by injected electric current...
- Noise-Immune Cavity-Enhanced Optical-Heterodyne Molecular SpectroscopyNoise-Immune Cavity-Enhanced Optical-Heterodyne Molecular SpectroscopyNoise-immune cavity-enhanced optical-heterodyne molecular spectroscopy is an ultra-sensitive laser-based absorption technique that utilizes laser light to assess the concentration or the amount of a species in gas phase by absorption spectrometry .- Principles :The NICE-OHMS technique combines...
(NICE-OHMS) - Tunable Diode Laser Absorption Spectroscopy